Where could the thermocouple be placed so that the


In Problem 7.23, an anemometer design was explored, and the assumption was made that the strip temperature was uniform. This is a good assumption when the heat transfer coefficient is low or the strip thermal conductivity high, because then conduction within the strip redistributes the generated heat and makes the strip temperature uniform. However, as the heat transfer coefficient increases or strip thermal conductivity decreases, heat generated at a point in the strip leaves the surface in the vicinity of that point, and the thermal condition is closer to one of uniform surface heat flux.

(a) Develop the calibration equations for both the constant surface temperature and constant heat flux conditions, that is, find the equations that predict the velocity as a function of the power per unit strip width, P'(mW/mm), and the temperature measured at the trailing edge (as in Problem 7.23). Assume laminar flow conditions.

(b) If the true condition is uniform surface heat flux, but the uniform surface temperature calibration is used, what percentage error will be incurred in the velocity determination?

(c) Where could the thermocouple be placed so that the calibration is insensitive to whether the thermal condition is uniform surface temperature or uniform surface heat flux?

Request for Solution File

Ask an Expert for Answer!!
Mechanical Engineering: Where could the thermocouple be placed so that the
Reference No:- TGS02137456

Expected delivery within 24 Hours